A Visual inspection machine (For injectable products) is successfully qualified by KNAAP test study using Min / Max and intermediate size of vials during PQ.
But during the challenge test study, it is noted that 1 or 2 defective vials (defects included as challenge) are not detected by the system and are included in the good vials. The acceptance criteria for this challenge test is 100% i.e. all the rejected vials must be rejected and all good vials shall be accepted. The equipment manufacturer is unable to meet the acceptance criteria.
Risk Assessment : This could be a potential risk as any such defect, especially defect like particulate matter could be a risk to the patient.
Queries :
As 100 % detection of particulate matter in the vials is improbable, can we relax the acceptance criteria for the challenge test study as NLT 98 % instead of 100 %. (Considering an AQL testing by production and QA personnel will follow).
What is the industry practice for the acceptance criteria for qualification of automatic visual inspection machine?
What is the challenge test and acceptance criteria followed prior to start of each commercial batch in the industry?