our product has 3.0% lod (specification limit of LOD is 5.0%w/w).before cheking the MR , any necessacity to dry the sample.( i mean with 3% LOD content sample or dried sample).please give sujjetion.
vasu
sir,
our product has 3.0% lod (specification limit of LOD is 5.0%w/w).before cheking the MR , any necessacity to dry the sample.( i mean with 3% LOD content sample or dried sample).please give sujjetion.
vasu
[quote=srinivas2202]sir,
our product has 3.0% lod (specification limit of LOD is 5.0%w/w).before cheking the MR , any necessacity to dry the sample.( i mean with 3% LOD content sample or dried sample).please give sujjetion.
vasu[/quote]
Dear Vasu
MR is done with drying, for the materials which have high moisture, It should be specified in the monograph.
[quote=srinivas2202]sir,
our product has 3.0% lod (specification limit of LOD is 5.0%w/w).before cheking the MR , any necessacity to dry the sample.( i mean with 3% LOD content sample or dried sample).please give sujjetion.
vasu[/quote]
Dear srinivas DRIED SAMPLE SHOULD BE TAKEN FOR MELTING RANGE UNLESS IT IS ESTABLISHED ITS FORM DOES NOT CHANGE