Flow Cell Change - Validation Impact

Hello Experts,
HPLC was qualified with a flow cell designed for higher pressure. Now user wants to use a different flow cell and column to have lower internal pressure.
To what extent the OQ must be repeated is the question.
Please suggest.
US FDA

1 Like

Change the flow cell. Then calibrate the HPLC as per your SOP. This should include a section on wavelength accuracy.