How come the acceptance limit 0.013 bar/ 10 minutes has fixed?..Is there any specific reason?..What is the basis for that limit and how that value has come?..Plz let me know…
Thanks in advance.
How come the acceptance limit 0.013 bar/ 10 minutes has fixed?..Is there any specific reason?..What is the basis for that limit and how that value has come?..Plz let me know…
Thanks in advance.
I do not know where didi they come but this acceptance limit has been given in HTM 2010 & PDA Technical monograph 1 & EN 285 which is used for the autoclave validation
Leak test limit is 0.013 bar/min. or 1 mmHg/ min. as per above guidelines
[quote=mahaksinghchaudhary]I do not know where didi they come but this acceptance limit has been given in HTM 2010 & PDA Technical monograph 1 & EN 285 which is used for the autoclave validation
Leak test limit is 0.013 bar/min. or 1 mmHg/ min. as per above guidelines[/quote]
Sorry, is it 0,0013 bar/min? (0,0013bar = 1mmHg)