What is acceptance criteria for Method Transfter

Hi,

What is the acceptance criteria (%RSD or other parameter) for the following in method transfer

Assay by High Performance liquid Chromatography

Assay by Potentiometry

Impurities by High Performance liquid Chromatography

Regards
Srikanth

[quote=snagabiru]Hi,

What is the acceptance criteria (%RSD or other parameter) for the following in method transfer

Assay by High Performance liquid Chromatography
2%.
Assay by Potentiometry
2%
Impurities by High Performance liquid Chromatography
10%

Things to remember.

when low level limits are to be compared.

you have to justify statistically.
[COLOR="#008000"]Eg: 0.1% level , you can fix 0.03 is the difference
Above 0.1% level limits, you can fix 0.05 difference.[/color]

Regards,
Bujji Kanchi.

Thanks for Information

[quote=bujjikanchi][quote=snagabiru]Hi,

What is the acceptance criteria (%RSD or other parameter) for the following in method transfer

Assay by High Performance liquid Chromatography
2%.
[/quote]

But Intermediate Precision in the original laboratory has the limit of 3% RSD. So why would this be tighter at 2% than is required for the original lab?

[attach]502[/attach][attach]502[/attach]

please check the reference guideline by WHO.

Table 1 of the attachment clearly saying that assay would differ by 2%.

You may keep 3%, Since your Intermediate precision from Originating lab have the limit 3%.

do you find practically the precision or intermediate precision in case of assay deviated more than 2% RSD any where.

Regards,
Bujji Kanchi.


Method Transfer WHO.pdf (153.2 KB)

We’re under FDA, not WHO. That’s an ongoing issue, everplace is different, and not so spelled out in black and white.

Intermediate precision, % RSD is reported.
Method transfer,% difference is reported.

There is a difference in reporting style of acceptance criteria to one another.

However, USP or FDA does not have any related guidance for this acceptance criteria as far as I know.

We have submitted many ANDAs(USFDA) by adopting WHO acceptance criteria and got the approval for the same.

USP<1224> General chapter “Analytical method transfer” is not addressed any acceptance criteria.

But I am helpless to your query exactly.

Best Regards,
Bujji kanchi.
+91-8008889185.

Dear All,

Method Transfer acceptance criteria given below:

For Assay methods:

  1. System suitability should meet as per methodology
  2. % Assay results shold meet as per Specification
  3. %RSD of assay results of each laboratory and two laboratories NMT 2.0.
  4. The absolute difference b/n the laboratories NMT 1.0.

For Impurities:

  1. System suitability should meet as per methodology
  2. % impurity results shold meet as per Specification
  3. %RSD of each impurity results of each laboratory and two laboratories NMT 10.0.
  4. The absolute difference b/n the laboratories based on the impurity limit. impurities below 0.05% comparison is not required.

Regards,
Srinivas.