Hello. I have a doubt. I used atomic absorption.
I read 3 times 3 standard.
One of the readings in blank (for temporary instrument error)
Can I use the other reading for calculation or I have repeat all over again?
Thanks
Hello. I have a doubt. I used atomic absorption.
I read 3 times 3 standard.
One of the readings in blank (for temporary instrument error)
Can I use the other reading for calculation or I have repeat all over again?
Thanks
It is normal to zero absorbance on the blank. You will have to have 3 standards at 3 concentrations (not just 2) since the curve is Lagrangian and not linear. Be sure your sample is in the range.
Ok the result of Absorbance is:
0.100
0.101
0.103
Mean 0.102
0.202
0.001 (!)
0.203
Mean 0.202
0.303
0.300
0.300
Mean 0.301
It is permissible?
The result of Sample is: 0.101
Not really. It depends on the sample. If it is a pharmaceutical you must perform a ‘laboratory investigation report and a root cause analysis’. I presume the standard was not added to the container. Then reanalyze the sample.
For an environmental sample follow EPA procedures. (It’s been a long time for me).